Posted: Wed Jan 21, 2009 4:13 pm Post subject: E-31 E-34 Internal Voltage error |
 |
|
| Getting this error on a 5K. The developer was redone long before this error started. We've replaced all the wires. Appreciate any ideas. |
|
| |
|
|
|
 |
 |
 |
 |
 |
phan6622 doctor
Joined: 09 Feb 2007 Posts: 750 Location: Midwest  |
Posted: Wed Jan 21, 2009 6:48 pm Post subject: |
 |
|
| The front panel it will tell you which e-3X error is occuring. Then you can trouble shoot that particular one. use the search on this forum, you will find answers to each of those issues. |
|
| |
|
|
|
 |
 |
 |
 |
 |
doug s medical school grad

Joined: 27 Feb 2007 Posts: 60 Location: SW  |
Posted: Wed Jan 21, 2009 8:10 pm Post subject: |
 |
|
| Yeah, I think both seem to suggest wires, but they're new. |
|
| |
|
|
|
 |
 |
 |
 |
 |
Charlie Brown doctor

Joined: 09 Aug 2005 Posts: 434 Location: Nashville Tennessee  |
Posted: Wed Jan 21, 2009 9:06 pm Post subject: E-? |
 |
|
The main question is which code are you getting I have a few e-51-1 and found toner buildup under the blocks and once I clean them problem and today on a 6000 that has been having radom 51-1 found a fine crack on rh head
I'll keep my Guns, Freedom, & Money ...... YOU CAN KEEP THE "CHANGE!" |
|
| |
|
|
|
 |
 |
 |
 |
 |
9lives intern

Joined: 05 Jun 2006 Posts: 107
 |
Posted: Thu Jan 22, 2009 10:33 am Post subject: |
 |
|
Did you replace the wires before the error began, or are they new since the error began? Do you string your own wires or use pre-strung? Could it be possible that a wire has a "pig-tail"? Even on pre-strung wires I have seen pig-tails before. _________________ There is nothing more freeing than leaving a "perfectly good airplane" at 10,000 ft, landing safely, and packing up to go do it again! |
|
| |
|
|
|
 |
 |
 |
 |
 |
doug s medical school grad

Joined: 27 Feb 2007 Posts: 60 Location: SW  |
Posted: Thu Jan 22, 2009 11:55 am Post subject: |
 |
|
| The error has been coming up way too frequently for a couple of months. It doesn't seem to matter if the wires are new or not. I always string my own wires, and I'm pretty sure it's not a pig tail issue. It almost seems like whatever senses the error is too sensitive. If I turn the machine off and on a few times, I can get it to power up and it doesn't seem to have any image problems. Almost seems like it's detecting every toner spec on the tr/sep wires. |
|
| |
|
|
|
 |
 |
 |
 |
 |
KIPDOCTOR Site Admin

Joined: 31 Dec 1969 Posts: 1402 Location: Boston Area  |
Posted: Thu Jan 22, 2009 3:38 pm Post subject: |
 |
|
This was originally posted in the 2000/6000 forums, but it is applicable to the 5000 as well.
Posted: 12 Jan 2005 08:45 am Post subject: Positive fix for E 51 issues
--------------------------------------------------------------------------------
OK, I am about to give you some that "unauthorized and unofficial" information right here, right now.
If you are sick and tired of E51 codes for what seems like no reason, read on.
It is my feeling the KIP 2000 and the KIP 6000 monitors the overcurrent on the chargers a little too closely. Sometimes just a spec of dust or toner can throw it into an E 51. If there was no monitoring circuit, the impurity would have just burned off and the machine would still be running. So what to do? Let's disable the monitoring circuit. Real simple operation.
On the bias and charger board in the 6000 and the 2000 (same board)
there is a connector CN 302. Here is the monitoring pin out for that connector.
PIN 6 1st Charge
PIN 7 Transfer
PIN 8 Separation
So, if you want to stop the machine from monitoring these signals, simply cut each wire and tape the end so they don't touch anything. The machine runs fine without them. Keep in mind, if you really do have a 1st charge shorting, you will still get a black print. It does not stop the charger from shorting, just ignores the overcurrent signal.
Please use this information at your own risk. If you don't understand what you are doing, please seek competent technical advice. I take no responsibility for the information contained above. ( legal crap). This works. _________________ What we have gained in technology, we have lost in humanity. |
|
| |
|
|
|
 |
 |
 |
 |
 |
doug s medical school grad

Joined: 27 Feb 2007 Posts: 60 Location: SW  |
Posted: Thu Jan 22, 2009 5:30 pm Post subject: |
 |
|
| Nice! thanks! |
|
| |
|
|
|
 |
 |
 |
 |
 |
Shyloh medical school grad
Joined: 09 Oct 2008 Posts: 97
 |
Posted: Fri Jan 23, 2009 4:16 pm Post subject: |
 |
|
Just a tip, but the exact error your getting would be extremely helpful. _________________ Can we fix it! |
|
| |
|
|
|
 |
 |
 |
 |
 |
doug s medical school grad

Joined: 27 Feb 2007 Posts: 60 Location: SW  |
Posted: Fri Jan 23, 2009 4:55 pm Post subject: |
 |
|
| Thanks! I realize that. We just replace wires too often on this one. Next call for this error I'll verify which exact error we've got. I've replaced the wires myself, and as I stated, still received the error for a couple of reboots, and returned for the same call the next day. The machine is running right now and I won't make any changes to the board to bypass the errors until I'm certain that's our last resort. I do as always, however, appreciate all input. |
|
| |
|
|
|
 |
 |
 |
 |
 |
colt-a3 dreaming of medical school

Joined: 30 Apr 2009 Posts: 4 Location: portland OR.  |
Posted: Fri May 22, 2009 1:48 pm Post subject: internal voltage error e-031 e-034 on kip 7000 |
 |
|
i am getting this constantly on a 7000 somtimes it is the charge and somtimes t/d, i replace endblocks and it is good for a month and then goes constant error again, will the wire snip work for a 7000??? _________________ You can unlock any door, if you only have the key........or a Twelve Gauge Shotgun! |
|
| |
|
|
|
 |
 |
 |
 |
 |
doug s medical school grad

Joined: 27 Feb 2007 Posts: 60 Location: SW  |
Posted: Fri May 22, 2009 2:38 pm Post subject: |
 |
|
| I think it is the same on the 7000. I never actually cut the wires. I explained the wire cut with my customer and he was willing to entertain the idea if the issue didn't resolve after replacing the whole Tr/Sep unit. However, after replacing the unit, the problem has not yet returned. |
|
| |
|
|
|
 |
 |
 |
 |
 |
colt-a3 dreaming of medical school

Joined: 30 Apr 2009 Posts: 4 Location: portland OR.  |
Posted: Fri May 22, 2009 2:41 pm Post subject: |
 |
|
i was thinking to just remove the wire from its holder rather than cut it to see if that works, they have no image problems... _________________ You can unlock any door, if you only have the key........or a Twelve Gauge Shotgun! |
|
| |
|
|
|
 |
 |
 |
 |
 |
doug s medical school grad

Joined: 27 Feb 2007 Posts: 60 Location: SW  |
Posted: Fri May 22, 2009 2:47 pm Post subject: |
 |
|
| I only have 2 7000s I service. I'm not positive if the wires are the same as the other machines, but I would imagine they are. Maybe someone else can confirm this? |
|
| |
|
|
|
 |
 |
 |
 |
 |
|
 |
 |
 |
 |