Posted: Mon Sep 15, 2014 1:24 pm Post subject: error code 06508 |
 |
|
Hey everybody, what is an 06508 and is the same for a 9800 and TDS800? If not, what is the difference?
Thanks,
David |
|
| |
|
|
|
 |
 |
 |
 |
 |
mark in vegas doctor

Joined: 20 Oct 2010 Posts: 1852 Location: Las Vegas, NV  |
Posted: Mon Sep 15, 2014 2:00 pm Post subject: |
 |
|
06508 MRE: ~Power-up: run CTFPMO not allowed~
06701 ORE: ~Paper too late at CTFOUTSE~
06-4-04
CTFPCLEHE
06-9-02
CTFPCLE-
HETMPSE
rising
Possible cause Remark
ON No Connection SSR PBA ->
CTFPCLEHE
SSR PBA defect
CTFPCLEHE / CTFPCLE-
HETEMPERSW defect
Rctfpclehe = 44
Ohm
Circuit breaker off
06-4-06
CTFSFVA
CONDRI-
OUTDIAG
2nd trans-
fer step con-
tinuously on
Possible cause Remark
OFF ON Yes DRIVERS PBA defect
Connection CTFSVA
-> DRIVERS PBA
CTFSFVA defect
OFF Yes CTFSFVA defect
CTFSFCYLL/R
mechanically defect
Don't have 800 TSM _________________ Regards,
Mark
RVN 68-69
It's so nice to be insane
No one asks you to explain |
|
| |
|
|
|
 |
 |
 |
 |
 |
jumeda33 intern
Joined: 20 Feb 2011 Posts: 146
 |
Posted: Mon Sep 15, 2014 4:14 pm Post subject: |
 |
|
| Thanks Mark. Any idea where to look or why it has a power up problem even though sometimes it will power up but most of the time it tries to come up but within 15 seconds or so it shuts back down. |
|
| |
|
|
|
 |
 |
 |
 |
 |
mark in vegas doctor

Joined: 20 Oct 2010 Posts: 1852 Location: Las Vegas, NV  |
Posted: Mon Sep 15, 2014 4:22 pm Post subject: |
 |
|
06-9-03 CTFPCLEHE-
TEMPSE
~CTFP CLeaner Heater TEMPSE (adc)~
(value A/D range 0 - 1023)
This is the cleaner roll heater sensor. I'd Take a look at anything to do with that heater.
Also look at the heater connector located on the frame below the motor plate towards the left as you face it. If it's burnt you'll need to replace pins and connectors. _________________ Regards,
Mark
RVN 68-69
It's so nice to be insane
No one asks you to explain |
|
| |
|
|
|
 |
 |
 |
 |
 |
planit medical school freshman
Joined: 28 Oct 2011 Posts: 19
 |
Posted: Mon Sep 15, 2014 5:55 pm Post subject: |
 |
|
On my TSM, 06508 says that, Power-Up Second Pinch Not Open.
Points to the Pneumatic Valves not opening and closing correctly.
You can run SDS test 066005 to check them.
Check compressor and lines.
All else fails, the knowledge base says that the control unit was faulty in some cases. |
|
| |
|
|
|
 |
 |
 |
 |
 |
jumeda33 intern
Joined: 20 Feb 2011 Posts: 146
 |
Posted: Mon Sep 15, 2014 6:26 pm Post subject: |
 |
|
Thanks Planit. I really appreciate that. So maybe I have an air leak somewhere or maybe the control unit is bad, correct? Also, is second pinch the big cylinders or the small ones.
Thanks,
David |
|
| |
|
|
|
 |
 |
 |
 |
 |
planit medical school freshman
Joined: 28 Oct 2011 Posts: 19
 |
Posted: Tue Sep 30, 2014 3:53 pm Post subject: |
 |
|
The second pinch involves the large cylinders. It occurs when the paper is between the fuser roller and the pressure roller.
The reason why its important to do a Torque adjustment test is to ensure the speed of the pressure roller and the fuser roller is correct, at that stage.
Here is a description of what the second pinch does.
The second transfer step uses heat an pressure to get a correct transfer and fusing of the toner on the print material. The print material is fed between the hard pressure roller and the soft fuser roller. The pressure roller and the fuser roller are pressed together using the 2 large pneumatic cylinders. This process is called the second pinch. The preheated print material is fed from the preheater unit into the second Pinch. The toner is transferred to the print material. |
|
| |
|
|
|
 |
 |
 |
 |
 |
|
 |
 |
 |
 |