buddha medical school freshman
Joined: 19 Oct 2009 Posts: 12 Location: Augusta GA  |
Posted: Mon Oct 26, 2009 11:53 am Post subject: Kip 9000 E21 and door open error |
 |
|
For a week now we have been trading parts in and out of our kip 9000 for an E21 error. I'm at my wits end fellas. I've replaced
1st DC Power Supply
2nd Thermistor
3rd The Other Thermistor
4th AC Circuit Board 2 Assy
5th Main Control Board Assy
We have replaced everything according to the manual that would or could cause this error and even some parts recommended by our daddy. The machine will not allow the mask due to the door open error. If I wasn't bald i would be now. Has any of y'all experienced this type of trouble? Not really even frustrated any longer just broken in spirit. Any help or suggestions would be greatly appreciated. Thanks in advance.
Buddha. |
|
| |
|
|
|
 |
 |
 |
 |
 |
cjwilt doctor

Joined: 24 Jan 2007 Posts: 1041
 |
Posted: Mon Oct 26, 2009 12:13 pm Post subject: |
 |
|
| Have you checked the fuser lamps yet? |
|
| |
|
|
|
 |
 |
 |
 |
 |
caltech doctor
Joined: 24 Feb 2006 Posts: 262 Location: Northern California  |
Posted: Mon Oct 26, 2009 12:41 pm Post subject: |
 |
|
| Check DCP-1 24 volt power supply, inside right door behind black cover. |
|
| |
|
|
|
 |
 |
 |
 |
 |
buddha medical school freshman
Joined: 19 Oct 2009 Posts: 12 Location: Augusta GA  |
Posted: Mon Oct 26, 2009 12:52 pm Post subject: |
 |
|
| lamps are good. |
|
| |
|
|
|
 |
 |
 |
 |
 |
buddha medical school freshman
Joined: 19 Oct 2009 Posts: 12 Location: Augusta GA  |
Posted: Mon Oct 26, 2009 2:13 pm Post subject: |
 |
|
Check DCP-1 24 volt power supply, inside right door behind black cover.
Checked it out and we've got 120v coming in from the wall but nothing coming out...
We will get that board |
|
| |
|
|
|
 |
 |
 |
 |
 |
caltech doctor
Joined: 24 Feb 2006 Posts: 262 Location: Northern California  |
Posted: Mon Oct 26, 2009 5:13 pm Post subject: |
 |
|
| You should have 240 coming from the wall not 120. You should check your wall voltage, if it is lower than 208 VAC you should install a buck/boost transformer. Low voltage is what cause power supplies to overheat and blow. |
|
| |
|
|
|
 |
 |
 |
 |
 |
cjwilt doctor

Joined: 24 Jan 2007 Posts: 1041
 |
Posted: Mon Oct 26, 2009 5:42 pm Post subject: |
 |
|
| Maybe they check 120 on each leg, but good point on the buck boost transformer. |
|
| |
|
|
|
 |
 |
 |
 |
 |
gath99 intern
Joined: 24 Jul 2007 Posts: 125
 |
Posted: Wed Dec 02, 2009 11:15 am Post subject: |
 |
|
Update on this problem
Turned out to be a DC power supply board. Part # 9000500225, on the lower right side behind a panel below the door. |
|
| |
|
|
|
 |
 |
 |
 |
 |
lead guitarist intern

Joined: 03 Apr 2007 Posts: 123 Location: Texas  |
Posted: Wed Aug 14, 2013 3:11 pm Post subject: |
 |
|
YEP 24V supply was my trouble as well, seems whom ever wrote the book at kip never worked in the field!!!! This is never mentioned in the very short FIP, for those old Xeroids you know what that means!!
Peace,
Lead Guiterist.
John B. _________________ If it's too LOUD get some plugs!! |
|
| |
|
|
|
 |
 |
 |
 |
 |
|
 |
 |
 |
 |