goodman medical school freshman
Joined: 07 Sep 2009 Posts: 22
 |
Posted: Wed May 04, 2011 8:29 am Post subject: RADF ERROR |
 |
|
| If somebody see this RADF error from KIP 7900.? What this error means? I already checked the manual i found nothing. Can you please help me... |
|
| |
|
|
|
 |
 |
 |
 |
 |
cjwilt doctor

Joined: 24 Jan 2007 Posts: 1041
 |
Posted: Fri May 06, 2011 9:07 am Post subject: |
 |
|
Check out Tech tip TIP-K120-7900-012 E-51 HV Error Indication.pdf
You have to log into service and find out what is causing this. It's probably an E51 code. |
|
| |
|
|
|
 |
 |
 |
 |
 |
dlee doctor
Joined: 07 Nov 2006 Posts: 1460 Location: E. Syracuse NY  |
Posted: Wed Jun 15, 2011 4:24 pm Post subject: |
 |
|
| I'm getting the same error on a brand new KIP7700 install. In the information screen it does not tell me which code. It tells me E-0051 Corona Charger Error. I'm thinking the charge wire or assy but I've checked and double checked it and cannot find anything. |
|
| |
|
|
|
 |
 |
 |
 |
 |
cjwilt doctor

Joined: 24 Jan 2007 Posts: 1041
 |
Posted: Wed Jun 15, 2011 4:36 pm Post subject: |
 |
|
| I have seen on a couple new 7700's the T/S end blocks had loose screws in the bottom of the assembly that holds them on. You might want to check that first. |
|
| |
|
|
|
 |
 |
 |
 |
 |
dlee doctor
Joined: 07 Nov 2006 Posts: 1460 Location: E. Syracuse NY  |
Posted: Thu Jun 16, 2011 10:01 am Post subject: |
 |
|
When I unplugged HV2, the print comes out but the image was very light. Then I took out the Transfer corona wire out of the T/S assy and reinstalled the assy. Same results, print came out but very light. New Tranfer corona wire didn't help. Swapped out the T/S assy and it works now. Brand new machine so I"m going to RA the assy.
Now, another issue to deal with. Every job submitted, first sheet out of the machine is a blank then the job starts to print. |
|
| |
|
|
|
 |
 |
 |
 |
 |
cjwilt doctor

Joined: 24 Jan 2007 Posts: 1041
 |
Posted: Thu Jun 16, 2011 10:29 am Post subject: |
 |
|
Is the print completely blank or can you see a very faint image?
If it's very faint it could still be a transfer issue. Maybe the on timing needs adjusted? |
|
| |
|
|
|
 |
 |
 |
 |
 |
dlee doctor
Joined: 07 Nov 2006 Posts: 1460 Location: E. Syracuse NY  |
Posted: Thu Jun 16, 2011 2:56 pm Post subject: |
 |
|
| Fixed after replacing the T/S assembly... |
|
| |
|
|
|
 |
 |
 |
 |
 |
PPCPete resident
Joined: 20 Feb 2005 Posts: 196 Location: Wisconsin  |
Posted: Mon Oct 10, 2011 1:43 pm Post subject: |
 |
|
Random Hi-volt errors can be caused by the Fuser IR lamp wires shorting to the chassis by the transport area where they plug in. Check and insulate these wires away from the sharp corners of the chasis. (Right hand side, near plug). _________________ PPCPete |
|
| |
|
|
|
 |
 |
 |
 |
 |
|
 |
 |
 |
 |