Posted: Wed Oct 29, 2008 1:16 pm Post subject: 3100 - E-000 on monitor / E004 in diagnostics - to clear |
 |
|
Have problem isolating, the issue. It gives the customer a E-000 error on the touchscreen, and when I go into servive mode to reset it the code is E-004 This is the second 3100 with this issue. Can reset error, and have on one replaced the DC control board, error came back. Replaced the two thermistors on one, and it has achieved a miraculous two weeks with out reoccurance. Has not resolved the second machines issues. I am apparently missing something, also just to note I have not had these issues on the 3000 models. I was wondering if anyone else has had this issue with the 3100 series.
Thanks in advance for any information on this.
Tom Williams  |
|
| |
|
|
|
 |
 |
 |
 |
 |
RMS intern

Joined: 05 Sep 2006 Posts: 111
 |
Posted: Wed Oct 29, 2008 4:17 pm Post subject: |
 |
|
I HAVE ONE 3100 THAT FROM TIME TO TIME WILL ERROR OUT WITH A E-004
COME TO FIND OUT IT WAS THE IN COMING POWER AT CUSTOMERS OFFICE. THEY DID NOT HAVE A DEDICATED LINE AND THE POWER VOLTAGES WERE VARIABLE. HOPE THIS HELPS YOU. |
|
| |
|
|
|
 |
 |
 |
 |
 |
dlee doctor
Joined: 07 Nov 2006 Posts: 1460 Location: E. Syracuse NY  |
Posted: Fri Jan 30, 2009 10:28 am Post subject: |
 |
|
OK, anyone figure what is going on here? I have one KIP3100 that had this occur 3 times this month already and 2 other KIP3100 that got this error within the past week. I am leaving shortly to another customer that is a couple of hours away for the same reason.
OK, Monday morning and the customer just called again. Same issue, E004, but I checked and double checked everything. Obviously, something is causing this. Unless I'm there when it errors out, I cannot seem to find anything. I'm going to replace the thermistor and thermistats. What is the board temp (service mode 2 - No. 03) on your machines? |
|
| |
|
|
|
 |
 |
 |
 |
 |
slyman doctor

Joined: 23 Apr 2008 Posts: 467 Location: Sherbrooke,Quebec  |
Posted: Mon Feb 02, 2009 5:08 pm Post subject: |
 |
|
Hey guys...I just came back from customer that doing the same code BUT ON A old 3000 !! The code appears intermittent and I already change the thermistor with same result but like I told you, it's intermittently. For me, it seem to came only in the morning at power on. When I was at customer, the machine was off and cold, put power on, reset code and machine was fine !!!
I only suspect one thing and please take a look, on my machine, I have 3 thermistors, at power on, TH1 is always 25 degrees UNDER the other ones but when the machine is on idle, TH1 is 5-10 degrees higher than the other two.
I tried to adjust the temp higher or lower with no good result...I put back it at default. I check all the bulletin/firmware...nothing about that.
I also check the voltage...ok at 121volt. ( stable )
If you have a idea...
Thanks |
|
| |
|
|
|
 |
 |
 |
 |
 |
LDavis medical school grad

Joined: 21 Apr 2007 Posts: 62 Location: Spokane  |
Posted: Thu Feb 26, 2009 1:56 pm Post subject: E-000 and E-004 |
 |
|
| Same here, The customer starts the machine up in the a.m. and the code appears once a week. Machine has 6K on meter and was installed 5 months ago. I go a reset and it is fine. |
|
| |
|
|
|
 |
 |
 |
 |
 |
dlee doctor
Joined: 07 Nov 2006 Posts: 1460 Location: E. Syracuse NY  |
Posted: Thu Feb 26, 2009 2:29 pm Post subject: |
 |
|
| Well, the 2 that I said had the problem once, has not had the issue come back. The third one that was constantly having the issue (2 or 3 times a day), I changed out both thermistors and changed out the inner (center) thermistat. It has been close to 3 weeks now with no issue. I call the customer every week to check and also tell him to inform me as soon as it errors out again. Since he is so far away, I showed him how to get into clearing the code, but so far so good. |
|
| |
|
|
|
 |
 |
 |
 |
 |
dlee doctor
Joined: 07 Nov 2006 Posts: 1460 Location: E. Syracuse NY  |
Posted: Thu Feb 26, 2009 2:32 pm Post subject: |
 |
|
| OH, I picked the center thermistat because I mistakenly ordered only 1 thermistat from KIP and did an enimeniminimo. Figured I would order a couple more and if it did error out, I would change the outer, but it has been good. Knock on wood. |
|
| |
|
|
|
 |
 |
 |
 |
 |
eboucher resident

Joined: 27 May 2008 Posts: 198 Location: Nashua, NH  |
Posted: Thu Feb 26, 2009 3:42 pm Post subject: |
 |
|
Has anyone noticed the condition of the heat roller? I am having the same problem at 2100 square and the heat roller looks worn. Also on the non drive sire the plastic bushing the heat roller rides in is wearing away at the roller itself. I just replaced the thermistors because that's all I had and it came right up as opposed to a 10 minute wait with the old ones. _________________ I'll have a tech give you a call shortly. |
|
| |
|
|
|
 |
 |
 |
 |
 |
slyman doctor

Joined: 23 Apr 2008 Posts: 467 Location: Sherbrooke,Quebec  |
Posted: Thu Feb 26, 2009 7:24 pm Post subject: |
 |
|
Hey guys, I had reply from Kip support, they ask me to do one thing, there is a "low temp room" button in the configuration option, I put it on and no more calls from the customer since then !!!!
The kip support told me this function turn on the heating a few times in the night .
So far so good. |
|
| |
|
|
|
 |
 |
 |
 |
 |
LDavis medical school grad

Joined: 21 Apr 2007 Posts: 62 Location: Spokane  |
Posted: Thu Feb 26, 2009 7:36 pm Post subject: |
 |
|
4th service call to me, 2nd to KIP, here is what was suggested after I recorded the line voltage and drops.
1) Turn low room temp on
2) Clean and check Thermistors
3) Install an ESP or electronic surge protector
Here are the voltages FYI,
Power on- 114 vac
Warm Up- 117 vac
Printing- 116.8 vac
Idle- 122 vac
Idle/Fuser maintaining temp- 116 vac |
|
| |
|
|
|
 |
 |
 |
 |
 |
LDavis medical school grad

Joined: 21 Apr 2007 Posts: 62 Location: Spokane  |
Posted: Thu Feb 26, 2009 7:41 pm Post subject: Error code |
 |
|
| I think if the time was increased that the machine detects the fuser roll temperature this would probably fix it to, just my 2 cents. |
|
| |
|
|
|
 |
 |
 |
 |
 |
eboucher resident

Joined: 27 May 2008 Posts: 198 Location: Nashua, NH  |
Posted: Mon Mar 02, 2009 10:32 am Post subject: |
 |
|
Argh...Kip just suggests the most obvious fixes. I have all those settings in place. I have never experienced an issue with voltage drops. We ran with the low room temp for a few days and it didn't resolve the issue. We went in and replaced the thermistors and thermostats. We'll see what happens. _________________ I'll have a tech give you a call shortly. |
|
| |
|
|
|
 |
 |
 |
 |
 |
eboucher resident

Joined: 27 May 2008 Posts: 198 Location: Nashua, NH  |
Posted: Thu Apr 09, 2009 9:17 am Post subject: |
 |
|
For those who don't know there is a firmware update to change the detection time for this error. I assume everyone here knows about it but I though I'd throw it out there. _________________ I'll have a tech give you a call shortly. |
|
| |
|
|
|
 |
 |
 |
 |
 |
dlee doctor
Joined: 07 Nov 2006 Posts: 1460 Location: E. Syracuse NY  |
Posted: Thu Apr 09, 2009 3:42 pm Post subject: |
 |
|
| Yup, saw the bulletin. But ever since I replaced the thermistors and thermistat in the one that was throwing the errors twice a day, they have had no issue since. |
|
| |
|
|
|
 |
 |
 |
 |
 |
|
 |
 |
 |
 |