Posted: Thu Oct 04, 2007 1:17 pm Post subject: Blown fuses |
 |
|
| We are experiencing fuses being blown on our PC controller PCB A Assy at a very high rate. What are potential causes of this issue? Could a faulty power supply on the controller cause the fuse to blow on the board to the KIP8000 model device? We have taken the power cord and have it connected directly to a power supply other than using the power from the KIP8000. |
|
| |
|
|
|
 |
 |
 |
 |
 |
Chazbo doctor
Joined: 05 Jul 2005 Posts: 377
 |
Posted: Thu Oct 04, 2007 6:08 pm Post subject: |
 |
|
change the fuse on the PC controller PCB to a 3 amp fast blow ceramic fuse _________________ Ride To Live |
|
| |
|
|
|
 |
 |
 |
 |
 |
KIPDOCTOR Site Admin

Joined: 31 Dec 1969 Posts: 1402 Location: Boston Area  |
Posted: Thu Oct 04, 2007 6:12 pm Post subject: |
 |
|
Yes, 3 amp is what we use also. _________________ What we have gained in technology, we have lost in humanity. |
|
| |
|
|
|
 |
 |
 |
 |
 |
Jay intern

Joined: 17 Feb 2005 Posts: 135 Location: Pensacola, FL  |
Posted: Thu Oct 04, 2007 9:25 pm Post subject: |
 |
|
| We had this happen to 2 7000's we had and our 2 8000's now. The 3 amp fixed the problem. |
|
| |
|
|
|
 |
 |
 |
 |
 |
dlee doctor
Joined: 07 Nov 2006 Posts: 1460 Location: E. Syracuse NY  |
Posted: Mon Dec 03, 2007 3:12 pm Post subject: |
 |
|
| This morning a customer had a 8000 that lost power to the controller. There is no power coming from the printer into the controller. But, if I connect the controller to the outlet on top where the stacker plugs into, I have power. So I trace the wires and get lost in the confusion but I do believe that it is a fuse. Is this the ceramic fuse on the middle right of the machine on the backside. Actually I believe that it is the only removable fuse back there. Sorry, I am not there right now so I cannot remember which board its on and very difficult to find in the service manual. Also, I recently replaced a fuse on our KIP6000. I used a 3.15 Amp, 250 Watt Slow-Blow fuse, but it is clear, not ceramic. Is this OK? |
|
| |
|
|
|
 |
 |
 |
 |
 |
dlee doctor
Joined: 07 Nov 2006 Posts: 1460 Location: E. Syracuse NY  |
Posted: Mon Dec 03, 2007 5:12 pm Post subject: |
 |
|
| Machine is up and running with the new fuse. Thank you. |
|
| |
|
|
|
 |
 |
 |
 |
 |
|
 |
 |
 |
 |