Posted: Wed Feb 27, 2013 1:55 pm Post subject: 9800 running hot |
 |
|
Hey guys. I have a 9800 that seems real hot when I open the covers. I have a second one that does not seem to be as hot when I open the covers. It will code a 06701 and sometimes an 06101. It all seems to lead to CTF temp sensor. It seems like it keeps pointing me to the NCTS or as I call it, the heartbeat thing. It is running and doing the heartbeat sound but is it possible that the end of it is dirty from toner or something and not reading correctly? Are they able to be cleaned in any way? Or, am I completely in the wrong direction and need to look elsewhere in the machine?
Thanks for any help.
David |
|
| |
|
|
|
 |
 |
 |
 |
 |
mark in vegas doctor

Joined: 20 Oct 2010 Posts: 1852 Location: Las Vegas, NV  |
Posted: Wed Feb 27, 2013 6:01 pm Post subject: |
 |
|
Pull the drun fan and blow it out including the sensor in the duct. Is the fan even running? Could be one of the heater triacs starting to fail. Make sure the other end of the drum isn't blocked, I had a dummy in production leave a paper plastic bag hanging over the end of the machine blocking the intake, wasn't a pretty sight. _________________ Regards,
Mark
RVN 68-69
It's so nice to be insane
No one asks you to explain |
|
| |
|
|
|
 |
 |
 |
 |
 |
jumeda33 intern
Joined: 20 Feb 2011 Posts: 146
 |
Posted: Wed Feb 27, 2013 6:05 pm Post subject: |
 |
|
| I put my hand over the place where the fan blows out and it was blowing air. Not sure how strong it should be. Other side is open and unblocked. |
|
| |
|
|
|
 |
 |
 |
 |
 |
mark in vegas doctor

Joined: 20 Oct 2010 Posts: 1852 Location: Las Vegas, NV  |
Posted: Wed Feb 27, 2013 6:15 pm Post subject: |
 |
|
Turn off power and clean the end of the NCTS sensor probe with a damp q-tip. If you clean it with power on you stand a good chance of shorting the heater and taking out the NCTS; I'm sure you can guess how I know. _________________ Regards,
Mark
RVN 68-69
It's so nice to be insane
No one asks you to explain |
|
| |
|
|
|
 |
 |
 |
 |
 |
jumeda33 intern
Joined: 20 Feb 2011 Posts: 146
 |
Posted: Wed Feb 27, 2013 6:17 pm Post subject: |
 |
|
Great. Thanks. I will clean the fan and the sensor in the duct and then clean the NCTS probe. Always wondered if those could get dirty and not sense properly. I will keep you posted and yes, I had a feeling you might have cleaned one without unplugging.
David |
|
| |
|
|
|
 |
 |
 |
 |
 |
printertechrep medical school senior
Joined: 12 May 2011 Posts: 31
 |
Posted: Thu Feb 28, 2013 9:22 pm Post subject: |
 |
|
| Hi, 06101 is an overtemp error, normally a triac, there are two behind the compressor. You can reverse them they are the same and see if the error code changes to the preheater which would be a 14101. |
|
| |
|
|
|
 |
 |
 |
 |
 |
jumeda33 intern
Joined: 20 Feb 2011 Posts: 146
 |
Posted: Thu Feb 28, 2013 9:28 pm Post subject: |
 |
|
Thanks Print,
I will check that out. Also, thanks for the reply on my other post about the image coming off the edge of the page. I will investigate that as well. I really appreciate everyones input as it is getting harder and harder to keep these old machines running.
David |
|
| |
|
|
|
 |
 |
 |
 |
 |
|
 |
 |
 |
 |